- Posts: 42
- Joined: Thu Nov 17, 2011 2:41 pm
Advertisement
Agilent 5971 BFB Tuning Problem
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
7 posts
Page 1 of 1
I am running a 5971 MSD that has been out of service for awhile and we are trying to get running again. The entire system has been flsuhed, the source cleaned/trap changed etc. I am having trouble tuning the bfb. I can get the ion ratios where I want them (131 = 40% , 219 = 44% relative to 69) but all of the isotopes are missing (ALL of them) when I run my BFB. Any suggestions?
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
Is resolution OK?
Width of peak at half height?
Abs. abundance of 69amu?
Width of peak at half height?
Abs. abundance of 69amu?
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
What does the autotune set the multiplier voltage at ? You may need a new EM as they don't like to be exposed to air for very long.
-
- Posts: 42
- Joined: Thu Nov 17, 2011 2:41 pm
EM voltage is around 2450 (seems high)
Peak width of the tuning peaks is .52, .53, .50 and the width at half height for the BFB peak is .07s
The abs abundance for 69 is 384640
The chromatography is a bit noisy and my peak shapes are not great either. The BFB peak shows ~1500 abundance while the noise level is around 400
Peak width of the tuning peaks is .52, .53, .50 and the width at half height for the BFB peak is .07s
The abs abundance for 69 is 384640
The chromatography is a bit noisy and my peak shapes are not great either. The BFB peak shows ~1500 abundance while the noise level is around 400
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
You need new EM.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Are you shooting enough BFB? EPA 524 allows for 25 ng of BFB on column. How fresh is your BFB solution?
I also think you need a new EM.
I also think you need a new EM.
-
- Posts: 42
- Joined: Thu Nov 17, 2011 2:41 pm
working on getting a new EM now. Will report back once its installed, thanks guys.
7 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science