Page 1 of 1
Residual solvent in head space GC
Posted: Wed Dec 07, 2011 10:13 am
by ananthjkumar
In Toluene sample,which is the suitable solvent and incubation temparature.
Re: Residual solvent in head space GC
Posted: Thu Dec 08, 2011 12:37 am
by chromatographer1
A sample consisting of toluene? and you are looking for solvents within the toluene? or a sample containing toluene and you wish to separate the toluene from the sample?
best wishes,
Rod
Re: Residual solvent in head space GC
Posted: Thu Dec 08, 2011 3:53 am
by ananthjkumar
Sample containing toluene and recover the same using HS
Re: Residual solvent in head space GC
Posted: Thu Dec 08, 2011 4:49 am
by chromatographer1
If the sample is fully dissolved into a non-polar solvent then a less then .25mL sample solution should equilibrate in 15 min or less.
In water solution I would suggest that 8 minutes or less will be adequate.
I would use a temperature of 80C to 85C.
best wishes,