what is the most common reasons for clogging of ESI probe
Posted: Thu Dec 01, 2011 4:44 pm
by jiangds06
i am running a LC-MS instrument. One day ESI capillary got clogged. I just wonder LC column should be very good filter for these particles. why LC column did not stop these particles?
Re: what is the most common reasons for clogging of ESI prob
Posted: Mon Dec 05, 2011 1:21 pm
by lmh
anything that is soluble can get through the column; anything that is not volatile will dry out in the spray chamber and may block either the capillary linking the spray chamber to the vacuum part of the instrument, or occasionally the spray needle.
The usual suspects are salty buffers in the solvents, but repeated injections of concentrated samples can do the same thing.
Re: what is the most common reasons for clogging of ESI prob
Posted: Tue Dec 06, 2011 6:11 pm
by MMJ88
Seconding what lmh said. It's probably a salt in your solvents. I have encountered this on my instrument.
Re: what is the most common reasons for clogging of ESI prob
Posted: Wed Dec 07, 2011 1:29 pm
by jiangds06
But i do not salt in my buffer. only volatile buffer.
Seconding what lmh said. It's probably a salt in your solvents. I have encountered this on my instrument.