Acylation Derivatisation
Posted: Wed Apr 20, 2005 8:56 am
by Chris
I am performing an Acylation derivatisation of OH groups for an active at low levels. Some of the documentation has highlighted that a siliconization of the glassware is performed.
Is this the same as silanization? as to reduce contamination I am going to perform the derivatisation directly in the HPLC vial and then inject and need to know if silanized vials are critical.
Posted: Thu Apr 21, 2005 11:18 pm
by tom jupille
I'm pretty sure that "siliconization" is a synonym for "silylation". As to the necessity, I can't help there.
Posted: Fri Apr 22, 2005 9:11 am
by Chris
Thanks Tom
I tried replicate derivatisations in both Siliconised and non Siliconised values and the recoveries and %RSD were similar so it should be ok.