Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
I've read in several post across the forum that buffers concentration should be between 10 and 50 mM and that a good starting point is 25 mM.
These concentrations are referred to the full eluent or to the aqueous portion of it?
Means to prepare an eluent 25 mM Acetate buffer 50:50 MeOH/Water, Do I mix 500ml of MeOH and 500 ml of 25 mM Acetate buffer or 500 ml of MeOH and 500 ml of 50 mM Acetate buffer?
Usually (but, alas, not always!) buffer concentration and pH are stated for the aqueous portion of the mobile phase.

By the way, there is nothing magic about that 10-50 mM range nor about the 25 mM. Those just happen to be where most people work in reversed phase with UV detection. You need sufficient buffer to keep ionization of your analytes and residual silanols under control. Increasing the concentration past that point provides no chromatographic advantage, is tougher on the hardware, and increases the risk of precipitation problems.

In principle, you want to use the lowest concentration that gives adequate pH control. In practice, most people don't attempt to optimize; they just use a "safe" 25 mM and let it go at that.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I agree with you. Buffer conc. should be as low as possible.
But optimizing that conc. can be a very long process.
This is the reason I did the question, at least minimize the range of concentrations to test.
Is anywhere a systematic approach to do that optimization?
Thanks you
Start at 0 and do a series of runs increasing the concentration in geometric progression (e.g., 0, 1, 2, 4, 8, . . . mM, or 0, 1, 3, 10 . . .mM). The runs should use "real" samples or spiked matrix blanks.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Talking about real samples... as soon as I lowered the buffer concentration (Acetate pH4.6) to 10 mM. the result for my wine control sample lowered in sorbic content from 186 to 171 ppm (the certified value is 190)
I don't understand what happens.
I use typical conditions: 1 ml/min of 50:50 MeOH/buffer in a C18 250x 4.6 column at 30ºC. Readings at 235 nm.
The most incredible is that I shoot a 100 mg/l Standar before and a 50 mg/l later and recoveries for them were 98-102%. over several injections...
I want to add that peak shapes were perfect in every injection, even in wine
Sometimes subtle peak shape variations can change the integration. If this were my problem, I would try going up to 50 and see if the recovery increases or if it plateaus (I'm assuming the initial condition was 25 mM).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I will work on it. Next week I hope I will have the data.
Thank you again
8 posts Page 1 of 1

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry