krickos » Tue Nov 29, 2011 4:31 pm
I finally jacked up the loop and transfer line temps to 180°C. Precision for pyridine on the next attempt was 3.5%. It took about a day before the blanks were clean enough to run though. I have a couple addition questions to which I couldn't find the answers from google.
#1. I downloaded a copy of Agilent's steam cleaning procedure. It states that the HS transfer line needs to be disconnected before the procedure is followed, but it seems like it was written for use with a MS unit. I just have a regular 6890 agilent GC using a DB-Wax column. Will this column be damaged from the water vapor if I didn't disconnect the HS transfer line?
#2. If the transfer line is disconnected from the GC, will the GC sense a leak condition and fault out? If so how do I avoid this?
#3. Is it ok for the headspace oven temp to exceed one or more of the boiling points of the analytes to be tested? I'm unclear about this. In my assay there are 5 components who's boiling points range from ~80°C-115°C. What oven temp do you guys suggest? What is the rule of thumb?
It's tough being a novice and I appreciate the help. thx
Hi
Thinks its clear that you had residual DMSO residues in your system based on the number of blank runs, seen that before. To monitor steam clean effectivness I never used a WAX column more sensitive than others(always did dedicate for DMSO/DMAA/DMF use or water), typically used a like a DB-5 or similar that is less sensitive starting at 100°C in column oven. Even a retention gap would prob do.
1. Yes thats for MS detector. When I had a MSD to work with we always had a FID as well as 2nd detector so we just shifted detector to monitor steam clean effectivness.
2. I prefer to monitor the steamclean effectivness so not a problem if you keep it connected to FID.
3. Absolutely!. The point is that the gas sample should ideally NOT interact with anything in the sample path way until it reaches the column. A good rule is to kept needle/valve/loop at +10°C above the highest bp of analytes including the sample diluent (DMF/DMSO/DMAA....). PArticulary for HS like Agilent that has a valve/loop injection system.
Finally, it can be adviseble to consider regular steamcleaning based on your sample throughput as a preventive maintenance schedule.