Hello everyone
Today I encountered a problem while using the LC/MS/MS which I have not previously encountered. The system is an Agilent1100 coupled with a QTrap
Prior to every experiment, I do system validation with reserpine. However on doing that today, I did not observe any peak.Instead I see intermittent tiny peaks (of about 10-30cps) unlike my previous peak (of over 1000cps)Also, previously, I had just one peak at abt 25sec.However, today I saw about 5 little peaks at different times. I have done all the troubleshooting I can think of but no progress yet
Please has anyone encountered this before?How was it resolved????...or any recommendations/suggestions
I cant do further experiments because of this as my sample also is not giving my desired peak
Thank you
Check LC: pump flow, injector etc. (may be the vial is empty ;) )

Check MS: infuse the reserpine solution using a syringe pump (+ESI, Scan mode 300-650amu)
2 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry