- Posts: 4
- Joined: Fri Sep 30, 2011 5:18 pm
Thanks,
Lisa
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
The "No emission current" error usually means that you have a bad filament. Normally they are either working or not ,with no in between, so the fact you got it to tune after seeing that error is a little strange. I'm not overly familiar with the 5970 since the 5973 is the oldest I have personally worked with, but if it has a dual filament source, try switching to the opposite filament and see if the error goes away.Hi, all. I am working with a HP 5890 and receiving an error message. Upon sending method to system, system returns MS Error = 10 (which is "No emission Current" "Excessive Signal Noise") The acquisition log lists 3727 Error in SCCONT Routine. None of this means much to me - only been on GC-MS for a year w/ little theory training. I ran an autotune successfully after error message. The first time I got the error I retried a couple of times and the method ran. Now it is not. Anyone have any ideas of what the problem is or anything I can do to better diagnose it? Have a call in to service contract w/ Spectro.
Thanks,
Lisa
Hi, all. I am working with a HP 5890 and receiving an error message. Upon sending method to system, system returns MS Error = 10 (which is "No emission Current" "Excessive Signal Noise") The acquisition log lists 3727 Error in SCCONT Routine. None of this means much to me - only been on GC-MS for a year w/ little theory training. I ran an autotune successfully after error message. The first time I got the error I retried a couple of times and the method ran. Now it is not. Anyone have any ideas of what the problem is or anything I can do to better diagnose it? Have a call in to service contract w/ Spectro.
Thanks,
Lisa
The "No emission current" error usually means that you have a bad filament. Normally they are either working or not ,with no in between, so the fact you got it to tune after seeing that error is a little strange. I'm not overly familiar with the 5970 since the 5973 is the oldest I have personally worked with, but if it has a dual filament source, try switching to the opposite filament and see if the error goes away.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.