by
EK8 » Tue Oct 22, 2024 6:26 pm
"EK*: wrote: "Waters claims we can inject 0.1 to 100uL".
- That is not a claim. That is a fact. You can set the injector at those values.That is the Specification range only, It has nothing to do what we are discussing here. A volume Input value in the software is different than actual sample volume on the column. Depending on how the HPLC is setup (the tubing used, the loop size, the sample concentration, the column, the mobile phase, the injection volume and sampling rate etc) will be the determining factor in what the peak will look like (and your separation). For your injector, you can set any number into the software between 0.1 and 100 ul. *Most auto-injectors have a "useful: range which gives the most accuracy between 20 and 80ul (Or for manual injectors, by overfilling the loop 3 or more times for best results). Multiple stacked injections can also be used too. But you are not there yet. What we are concerned with is sample dilution and diffusion using loop sizes that are not optimized for the A/I. As noted, we are less concerned about accuracy in your case. More concerned that you seem fixated on the injector instead of using the stock setup and actually determining what the chromatography looks like using different sample concentrations. Start running some samples using your method to get the process started.
The problem is I called Waters and the technician said the sample will not be diluted in the loop, even when using a low injection volume in a large loop. How this works, I don't know, but it is contradictory to what you are saying. I need to find out who is being factual.
To add - why should I care if the sample is diluted with mobile phase in the loop, assuming when I collect fractions for bioactivity assessment, I reduce each collected fraction to the original volume of my injected sample?