Posted: Mon Nov 21, 2011 9:45 am
				
				Hi all,
I am just starting as an environmental science technician and we are running a brand new dionex 2100 RFIC to run analysis of environmental water samples.
I have run into a bit of a problem that hopefully one of you experts will be able to help me with: During the analysis of the water samples, I am getting a peak that corresponds to nitrite in the sample. However, cross validation against a colorimetric method (EPA 115) indicates that this peak is false.
I have run glassware blanks, reagent blanks, check standards and multiple injections to get to the bottom of what is happening and I am still confused. I have no doubt that the "nitrite" peak is real, as it turns up in samples but not in the blanks, we have cross-validated between the IC and the colorimetric method and determined that the IC and other method are comparable to 0.1 PPM, so we are certain that we are not seeing shadows.
The final thing that we can do is run an internal spike to see what happens, but I suspect that there might be something eluting at about the same time as I would expect to see nitrite. Has anyone got any suggestions or advice to help me determine what is going on? -I know that we have a meeting with the local EPA coming up and I really don't want to be in a position of not knowing why I am getting "nitrite" where there should be none.
The method that we are running is 33mM KOH at 1mL/min with a dionex AS18 column heated to 30degress and conductivity detection.
Thanks for the time.
James
			I am just starting as an environmental science technician and we are running a brand new dionex 2100 RFIC to run analysis of environmental water samples.
I have run into a bit of a problem that hopefully one of you experts will be able to help me with: During the analysis of the water samples, I am getting a peak that corresponds to nitrite in the sample. However, cross validation against a colorimetric method (EPA 115) indicates that this peak is false.
I have run glassware blanks, reagent blanks, check standards and multiple injections to get to the bottom of what is happening and I am still confused. I have no doubt that the "nitrite" peak is real, as it turns up in samples but not in the blanks, we have cross-validated between the IC and the colorimetric method and determined that the IC and other method are comparable to 0.1 PPM, so we are certain that we are not seeing shadows.
The final thing that we can do is run an internal spike to see what happens, but I suspect that there might be something eluting at about the same time as I would expect to see nitrite. Has anyone got any suggestions or advice to help me determine what is going on? -I know that we have a meeting with the local EPA coming up and I really don't want to be in a position of not knowing why I am getting "nitrite" where there should be none.
The method that we are running is 33mM KOH at 1mL/min with a dionex AS18 column heated to 30degress and conductivity detection.
Thanks for the time.
James