RS - Method validation for undiluted sample (injectables)
Posted: Fri Nov 18, 2011 10:00 am
by mylsamyp
Hi
I would like to know the procedure for RS method validation of one injectable product.
The method is directing to use undiluted sample (i.e. 1 mg/mL - as such sample) for RS method analysis. Injection volume is 100 µL.
I need to validate the method now. How can I spike the impurities to sample?
Can I go for concentrated formulation product and do 5 mL to 10 mL dilution in which I can spike the impurities?
Expecting someone to guide me in this...
Re: RS - Method validation for undiluted sample (injectables
Posted: Fri Nov 18, 2011 4:11 pm
by Consumer Products Guy
That sounds like a good approach.
Posted: Fri Nov 18, 2011 4:30 pm
by Rohit Hirpara
Dear
First you prepare concentrated stock solution of impurity which can you spike,and after spike as low as possible volume lke 0.1ml,0.2ml----0.5ml of impurity stock solution as per your requirement.I think less volume of impurity stock solution could not effect onquantification particularly in Related impurity method.
Posted: Sat Nov 19, 2011 7:38 am
by mylsamyp
Thanks for your quotes.
But in this case, how can I proceed for Forced degradation (acid, base & Peroxide) experiment?...