- Posts: 6
- Joined: Thu Oct 13, 2011 4:34 pm
I am trying to run a test for impurities using octanoic acid as an internal standard. I keep getting an RSD failure because the octanoic acid peak area keeps falling each injection. The method has been validated and run before but this problem keeps cropping up. I am wondering if anyone has encountered this and if they have any fixes that would not require a revalidation. I am using a liquid inject method on a shimadzu gc 2010 and a liner with deactivated wool. The column is an agilent EC-1000 (30 m 0.32 mm 0.25 micrometer phase thickness) with a 5 m guard column. The diluent for the method is denatured ethanol. Here are the method parameters:
Injection Port Temperature: 220°C
Split Ratio: 10:1
Detector Temperature: 250°C
Carrier Gas: Helium
Carrier Pressure: 4.4 psi (~30 kPa)
Injection Volume: 1 µL
Detector Gases: Air: 400 mL/min
Hydrogen: 47 mL/min
Run Time: 50 minutes
Column Oven Program: Initial Temperature: 130°C
Initial Time: 3.5 min
Rate 1: 3°C/min
Temperature 2: 200°C
Hold Time: 0 min
Rate 2: 25°C/min
Final Temperature: 225°C
Hold Time: 22.17 min
Column Pressure Program: Initial Pressure: 4.4 psi
Initial Time: 25.0 min
Rate 1: 1.5 psi/min
Final Pressure: 11.6 psi
Hold Time: 20.0 min
Any ideas would be greatly appreciated.