If you have a leak on the other side of the regulator or the valve (on the inside of the PAL) is open, the pressure will drop to near 0. Please make sure that you have no leak before and after the regulator and it would be best to turn off the PAL autosampler so that the internal valve is closed. Then regulate the pressure to about 0.5 - 1 bar. Turn the PAL back on and check that the pressure stays at that level. Lastly, when you flush the syringe (during the headspace cycle), you will see the pressure drop to near 0.
If you require new o-rings or regulator, feel free to contact us. The o-ring is P/N: ORing 3.68x1.78GA-65, the fitting on the regulator MM 09-01 and the entire regulator with fittings MM 03-01.