Empower 2 report chromatograms (FD) Energy---> millivolts

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hello everyone,

i`m new to this forum, so don`t be very angry if i ask a question which has been asked before or if it`s a silly one.

Here`s the question:

For hplc analysis i use the fluorescence detector W2475. In the instrument method it is checked, that the data mode is "ENERGY" (data units - Energy). In Empower 2 report i want to add a chromatogram where the Y axis would be mV, not energy units. Is it possible? The height of the peak in the tables is shown in μV, but i just can`t find how to plot mV against time.

Any help will be appreciated,

Vaidotas.
Within the report method, in the Chromatogram object's properties, first tab - you can specify whatever labels for axes you like. It can read Fortnights v. Furlongs if you so desire. The Y axis will tend to report your detector's output in terms of Absorbance Units no matter the label assigned.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry