HS-sample preparation problem.
Posted: Wed Nov 09, 2011 8:06 pm
I should spike the residual working standard into a saline solution for accuracy validation. Saline solutions were prepared by diluting the stock standard solution to the required concentrations (5 levels over the range 1 ppm to 120 ppm of 1-propanol and MEK).
The specifications of the residual solvents in the saline: 1-propanol (NMT 100 ppm) and MEK (NMT 35 ppm).
Preparation:
Stock standard solution 1: 50, 0 mg of 1-propanol / 50 ml H2O
Stock standard solution 2: 50, 0 mg of MEK / 50 ml H2O
Range 1: required conc.: 1ppm 1-propanol and 1 ppm MEK
A: 1,0 ml stock std 1+ 1,0 ml stock std. 2/ 100,0 ml H2O.
Add 20µl of A + 180µl saline to HS-vial 1.
Range 2: required conc.: 10ppm 1-propanol and 5 ppm MEK
B: 10,0 ml stock std 1+ 5,0 ml stock std. 2/ 100,0 ml H2O.
Add 20µl of B + 180µl saline to HS-vial 2.
Range 3: required conc.: 50ppm 1-propanol and 10 ppm MEK
C: 25,0 ml stock std 1+ 5,0 ml stock std. 2/ 50,0 ml H2O.
Add 20µl of C+ 180µl saline to HS-vial 3.
Range 4: required conc.: 100ppm 1-propanol and 35 ppm MEK
D: How can I prepare the solution which I later use 20µl for vial 4 ?
Add 20µl of D + 180µl saline to HS-vial 4.
Range 5: required conc.: 120ppm 1-propanol and 42 ppm MEK
E: same question like in D.
Add 20µl of E + 180µl saline to HS-vial 5.
If you have another suggestion, can you please describe the step by step procedure of how to prepare samples and do the calculations? That would be great.
Parallel I want today to test the solution of saline, how many concentrations of 1-propanol and MEK they have. The RSDs for peak area of 1-propanol from 3 samples were 16% (around 2,2 ppm) and MEK wasn’t present. But the RSD values of the standard solution (two analytes) were below 2%. I think it’s critical for this analysis because saline plays a main role in recovery. Should I heat saline solution? If yes how much degree and time should I use?
Thanks.
The specifications of the residual solvents in the saline: 1-propanol (NMT 100 ppm) and MEK (NMT 35 ppm).
Preparation:
Stock standard solution 1: 50, 0 mg of 1-propanol / 50 ml H2O
Stock standard solution 2: 50, 0 mg of MEK / 50 ml H2O
Range 1: required conc.: 1ppm 1-propanol and 1 ppm MEK
A: 1,0 ml stock std 1+ 1,0 ml stock std. 2/ 100,0 ml H2O.
Add 20µl of A + 180µl saline to HS-vial 1.
Range 2: required conc.: 10ppm 1-propanol and 5 ppm MEK
B: 10,0 ml stock std 1+ 5,0 ml stock std. 2/ 100,0 ml H2O.
Add 20µl of B + 180µl saline to HS-vial 2.
Range 3: required conc.: 50ppm 1-propanol and 10 ppm MEK
C: 25,0 ml stock std 1+ 5,0 ml stock std. 2/ 50,0 ml H2O.
Add 20µl of C+ 180µl saline to HS-vial 3.
Range 4: required conc.: 100ppm 1-propanol and 35 ppm MEK
D: How can I prepare the solution which I later use 20µl for vial 4 ?
Add 20µl of D + 180µl saline to HS-vial 4.
Range 5: required conc.: 120ppm 1-propanol and 42 ppm MEK
E: same question like in D.
Add 20µl of E + 180µl saline to HS-vial 5.
If you have another suggestion, can you please describe the step by step procedure of how to prepare samples and do the calculations? That would be great.
Parallel I want today to test the solution of saline, how many concentrations of 1-propanol and MEK they have. The RSDs for peak area of 1-propanol from 3 samples were 16% (around 2,2 ppm) and MEK wasn’t present. But the RSD values of the standard solution (two analytes) were below 2%. I think it’s critical for this analysis because saline plays a main role in recovery. Should I heat saline solution? If yes how much degree and time should I use?
Thanks.