spectra system P4000 pump solvent leakage

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
whenever I start the hplc and begin running my method, the pump head begins to leak. maintenance was done on the pump recently, cleaning and replacing seals and pistons, and ever since then it leaks whenever I run my method. I removed the pump head and piston holder and made sure it was put together correctly and closed tight but that didn't make a difference. Any suggestions would be appreciated.
You've already checked what would have been my first and second suggestions:
- are the seals installed correctly and
- is the piston scored.

You may have just gotten a bad set of seals (or inadvertently damaged them during installation).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I have a few extra seals so maybe I can replace them and see if that helps. Thanks.
3 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry