Problems in MRM mode

Posted: Tue Nov 08, 2011 8:44 am
by danielamecro
Hi all,
I have problems with my GC-MS triple Quad system.I have change the argon tank and i loss the sensitivity from run to run.Next day,the first analisys is good but the areas from pesticides decrase gradual for 10 replicates.I clean Q1 and Q2 but the problem is still theare.
In SIS mode all is OK.For exemple:HCB,the ion 284 in Q1 have the same sensitivity for 10 replicates.In MRM mode,decrease the area from 2MG counts to 100 kcounts.
What can i do?
Please advice!
Thanks

Re: Problems in MRM mode

Posted: Mon Nov 14, 2011 10:44 pm
by Alp
Leak?
Did you turn the tank on (valve)?

Re: Problems in MRM mode

Posted: Tue Nov 15, 2011 4:26 pm
by sav
Are you purge gas line (argon) before installation?
Purity of argon?

Re: Problems in MRM mode

Posted: Wed Nov 16, 2011 6:32 am
by danielamecro
Are you purge gas line (argon) before installation?
Purity of argon?
I purged the line but unable to check the purity.Argon is 5.0
I changed the forline pump,i assumed it could not evacuate argon in Q2.The next step is to change the tank.
Thanks
Daniela

Re: Problems in MRM mode

Posted: Wed Nov 16, 2011 10:41 pm
by sav
I changed the forline pump,i assumed it could not evacuate argon in Q2.
Vacuum/pressure in Q2 - ОК?
The next step is to change the tank.
Advice: purge Q2 with argon (12-48hours; just don't close collision gas valve) before the next step. This may help you.

Best wishes, Alex.

Re: Problems in MRM mode

Posted: Thu Nov 17, 2011 7:13 am
by danielamecro
vacuum in Q2 is OK
i appreciate your advice
Thenk you!