Validation of thermal conductivity detector (TCD)

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all,

My question is bit open ended.

I have to validate a TCD which one of my friends has developed. I have chosen air as my first analyte. It contains about 0.04% of CO2 which I am trying to detect. I am using Molecular Sieve and Porapak-Q packed columns with H2 as carrier. I have kept the carrier flow at 30ml/min, just to make sure that I do not burn my friend's first creation :). All heated zones of the GC, e.g. injector, oven, and the detector block, are at room temperature (~25C).

I am expecting you to report the best result that you have seen for atmospheric CO2 detection using a TCD GC. Also please give other parameters, e.g. injected volume, detector current, and column.

Thanks,
Krish
so not use sieve unless you are using 13x.

the q column should be fine.

inject 1mL or less.

good luck

rod
2 posts Page 1 of 1

Who is online

In total there are 167 users online :: 2 registered, 0 hidden and 165 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Bing [Bot] and 165 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry