Buffer concentration vs Retention Time
Posted: Fri Nov 04, 2011 10:35 am
by ether
Lately I try to separate multi component analyte using ion pair chromatography. By mistake, i make buffer concentration twice as it should be, but the result is facinating. The peak is separated better in bigger buffer concentration. I'm happy to see better result, but i'm wondered how it can be happen?
red chromatogram : 40 mM buffer concentration
blue chromatogram : 20 mM buffer concentration
thank u :)
Re: Buffer concentration vs Retention Time
Posted: Fri Nov 04, 2011 4:31 pm
by tom jupille
No great surprise if you visualize ion-pair chromatography as a two-stage process:
1. Some of the (charged) ion-pair reagent sticks to the column
2. Sample ions and buffer ions compete for access to those charged groups.
Essentially, you are doing ion-exchange chromatography. Increasing the ionic strength of the buffer decreases the retention of your analytes. In this case, different compounds changed by different amounts, which changed (improved) the selectivity.
Posted: Sat Nov 05, 2011 11:28 pm
by ether
thank u Tom for your explanation :)
Often in ion pair chromatography, i use buffer only to achieve solvent desired pH. I forget that the buffer ions can compete with the sample ions. Even though if I remember, I never thought it will give such a difference like that.
Can you explain me more about ion pairing, Tom? :D
For future development, how can i know the right ionic strength i should use to get good peak separation?
If you have peak like the blue one, how can you know that increasing ionic strength will improve the separation?
Posted: Sun Nov 06, 2011 11:29 pm
by tom jupille
To take the questions out of order :wink: :
You can't. What you have to do is to pick a reasonable value (usually in the 25 mM range) and then change it and see what happens.
Again, you can't. Increasing the ionic strength might have made it worse (in which case, decreasing it would probably improve the separation).
In our Advanced HPLC Method Development course. we spend at least a couple of hours on ion-pair chromatography (and even that manages to only skim the surface) so there's no way I can give a detailed explanation in a Forum post, but here's a grossly oversimplified version:
An ion-pair reagent is essentially a surfactant, with a hydrophobic tail and a (charged) hydrophilic head. The easiest way for me to visualize what happens is that the hydrophobic tail gets embedded in the hydrophobic stationary phase while the charged head stays on the surface in contact with the aqueous mobile phase. In effect, you end up with an ion-exchanger (the difference between this and a conventional "fixed site" ion exchanger is that in the latter, the ion exchange sites are permanently bound to the support, while in the ion-pair system, they are loosely bound by hydrophobicity). The advantage of ion-pair is that you have a "tunable" system: if you add lots of ion-pair reagent (or use a very hydrophobic ion-pair reagent), you will have a system that behaves primarily as an ion-exchanger. If you add only a little bit of ion-pair reagent (or use a not-very-hydrophobic ion pair reagent), it will behave primarily as a reverse-phase system -- and obviously, you can tweak it to be anything in between.
As with all reversed-phase separations, getting reasonable retention is (usually!) not that difficult. Most of the effort goes into selectivity (the relative spacing of the peaks). That is controlled by six parameters:
- solvent strength (% organic for an isocratic separation or %B/minute for a gradient).
- solvent type (the classical Acetonitrile/Methanol/THF triad); usually not that important in ion-pair unless the selectivity problem is with neutral compounds
- temperature
- pH
- buffer and/or additive concentration
- column chemistry.
We can make general predictions (e.g., adding more organic solvent makes everything come off sooner; adding more IP reagent makes ionics come off later and neutrals come off a bit sooner; changing pH can make ionics go either way but usually has little effect on neutrals; increasing the salt concentration makes ionics come off sooner but has little effect on neutrals; etc., etc.) but in any given case it comes down to experimentation: pick a parameter, make a big change, and see what happens.
Posted: Mon Nov 07, 2011 12:47 am
by ether
OK.. I can understand it now.. :D
Thank you so much Tom!
Posted: Tue Jan 16, 2018 11:02 am
by piyushatre
This is about ion pairing reagents. but what about other salts? for example if we incease KH2PO4 in my mobile phase what effect it will have on chromatography??
Posted: Tue Jan 16, 2018 11:18 am
by HPLC chemist
It means your mobile phase will have greater pH buffering capacity. While, in most cases 20 mM is strong enough to handle the API variable ionic forms it may also be consumed by the formulation. Thus, 'sometimes' 100 mM is used.
Your peaks 'may' also be sharper and thus your LOQ lower.