I had developed an UPLC method using a Acquity BEH RP-Shield column (1.7 µm, 2.1 mm x 100 mm) and and a weak acetonitrile gradient, going from 20 to 25% acetonitrile in 5 mM ammonium acetate pH 5.2 (over 10 minutes). Temperature: 50°C.
The problem with this method was that the Acquity column did not offer a stable separation. I could see a continuous loss of retention of the acidic compounds. The retention of the neutral compounds didn't change, which lead to co-elution after about 24 hours of usage of the column. The column seemed to have anion-exchange properties that was washed of during usage. This was confirmed by the late Uwe (thank you for all your help!). The official standpoint from Waters was that I had been very unlucky to see this effect (which I find strange, due to the loss of retention counting in minutes).
I was very glad too see that I could replace the Waters column with a Phenomenex Kinetex column (only a slight pH change to pH 4.6 needed). The Kinetex column also offered a fantastic improvement of plate count and peak symmetry, which was quite surprising. The problem was that this column was even more unstable, using the given conditions. After only a few hours use, some peaks even shifted elution order.
I have a lot of questions.....
- Has anyone seen the same thing?
- What are these anion-exchange groups (residues from column manufacturing acc. to Uwe)?
- Are there any UPLC columns that do not show this behaviour?
I am under quite a lot of pressure about this, since I was supposed to transfer this method to QC now. The method would simply not work in a QC environment. The obvious solution is to lower the pH so all components are neutral, but it doesn't work. All peak then elute at the same time. HELP!