Benzoyl Peroxides and C8 and C18 stationary phases
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
Did anyone had any kind of interaction between Benzoyl Peroxide flushing as a sample into C8 or C18 Stationary Phases? I mean that interaction collapses these phases?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I've only done the assay a few times, don't remember any HPLC issues. I think the only "issue" was obtaining a decent standard as the BP typically has high water content.
USP is about $300 for "everything", and typically only provides a miniscule amount.
USP is about $300 for "everything", and typically only provides a miniscule amount.
-
- Posts: 2
- Joined: Mon Sep 12, 2011 6:33 pm
My product is a gel with 0.1% Adapalene and 3.0% of Benzoyl Peroxide.
I´ve observed this problem ending one month validation for both actives, during this period I´ve done lots of sample inyection for both systems. And for adapalene valoration the concentration of Benzoyl Peroxide (BP) is about 0.3 mg/ml an in the sample of BP valoration is 0.03 mg/ml.
Both assays were preformed in the same column (Waters Symmetry C18 150x4.6 mm 5um) with different chromatographics systems, and the column flushing 0.3 mg/ml colapses more rapidly than the other.
I´ve a lot of experience with adapalane in the same gel base and it didn´t present a problem. The only addition of BP generate the problem.
I read that BP is used in the initial polimerization when stationery phase of HPLC columns is produced.
I´ve observed this problem ending one month validation for both actives, during this period I´ve done lots of sample inyection for both systems. And for adapalene valoration the concentration of Benzoyl Peroxide (BP) is about 0.3 mg/ml an in the sample of BP valoration is 0.03 mg/ml.
Both assays were preformed in the same column (Waters Symmetry C18 150x4.6 mm 5um) with different chromatographics systems, and the column flushing 0.3 mg/ml colapses more rapidly than the other.
I´ve a lot of experience with adapalane in the same gel base and it didn´t present a problem. The only addition of BP generate the problem.
I read that BP is used in the initial polimerization when stationery phase of HPLC columns is produced.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
As far as I know, benzoyl peroxide is not used to initiate any polymerization of C18 or C8 bonded phase packings, for the simple reason that no polymerization is involved. Most commonly, a reactive silane is used to attach the stationary phase to the silica support. BP *is* used as an initiator for polystyrene-divinylbenzene resins, but those are relatively uncommon for reversed phase work.
Which brings me to the next question: what do you mean by "collapse"? What symptoms/problems areyou seeing?
Which brings me to the next question: what do you mean by "collapse"? What symptoms/problems areyou seeing?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science