Failing specificity due to a gradient artifact

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,
We are testing the specificity of an old reverse phase method (gradient), we have found out that there is a gradient artifact showing right at the retention time of our peak of interest. We have considerably modified the gradient to move the system peak away from the compound of interst, to conlcusively determined that it is indeed a system's peak. Can anybody comment on how to approach specificitiy in a situation like this? This is my first time handling a specificity validation.
Kind Regards,
AR
If there is an artifact peak exactly at the retention time of the analyte, then the original method is, in fact, not specific for that analyte. Your modified method *is* specific for the analyte (obviously assuming you have checked the retention of other potential interferences).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 3 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry