- Posts: 6
- Joined: Wed Nov 16, 2011 11:41 am
I'm doing some research using C12 as a solvent and benzothiophene (BZT) as a sulfur compound. I have in my lab Agilent 6890N GC and 5973N MSD. The GC has two detectors FID for hydrocarbons and FPD for sulfur. The two columns are: Agilent 19091Z-205 350C max HP-1 capillary (50m*200microm*0.5 microm).
Right now I'm trying to do a calibration for FPD by injecting different concentrations of BZT 1-10wt/wt % dissolved in C12. First few runs with low concentration produced nice sharp and symmetric FPD peak. After that I'm getting broad peak and sometimes three broad peaks! The way I'm injecting is to my reactor which operates at 250C and the gas sample is sucked by vacuum to a 1000 sulfur coated sampling cylinder. Then through the transfer line to my GC. All temperatures for sample cylinder and transfer line are 250C. The gas used in reactor and sample cylinder is Argon so the sample is always diluted in Argon. Also, I noticed the peaks for C12 and BZT are attached in MSD i.e. not separated.
I'm listing my method parameters here:
For MSD: Acquistion mode scan, threshold 150 counts, sampling rate 2, mas range 5-500, scan/sec 2.97
For GC: mode split, gas He, temp 270C, prssure 31.1, split ratio 25. Ramping for the oven starts at 35C (set point) hold for 10 mins, then 8C/min to 70C no hold, then15C/min to 300C no hold.
I know my message is too long! but I appreciate help