the diferrence in area between the sample and the standard
Posted: Tue Oct 25, 2011 6:21 pm
i have a small problem, i am supposed to check the % of a pesticide in a product as labelled on the bottle of the pesticide, this is usually done by preparing 2 vials of a standard solution and sample solution to be tested both to 100 ppm, and then run them successively, if the % of the pesticide is correct then the areas should be similar or very near to each. sure their is a percent of error which is accepted by the HPLC and the product.
if the pesticide is viscous , the preparation of the sample is done by measuring weight and then do the needed calculations,
usually the preparations are done with ACN and the mobile phase is ACN:water 80:20.
i am using automated HPLC from agilent.
most times the results are perfect , but some times i get a perfect peaks of the standard and the sample , but their is much difference in the area and this makes the expected error more than the accepted.
i am sure from the preparation and the used method , and the product i am checking is a good quality one, so i know that it should pass and it is not that their is a cheat in the product.
the same used standard is used before for many times and the results were great , so i am sure also it is not the standard.
as i said the hplc is automated and the peaks are perfect , so their is nothing from overdose injection or contamination.
the thing that noticed was just the pressure was not constant all the time, it was decreasing with time from about 80 to 70 ,
i am really mixed , if anybody knows why it happens like that please help.
if you need more details i will be happy to paste it.