For Split and Splitless injector depths use the GC oven
Splitless is the distance from the oven edge to the top plate , Split is just the main oven width.
If you want the proper figures
PTV Injection (a)
Syringe Needle Length 51 mm
Column Insertion Depth 30 mm
Split Injection SSL (b)
Syringe Needle Length 51 mm
Column Insertion Depth 40 mm
Splitless Injection SSL (c)
Syringe Needle Length 80 mm
Column Insertion Depth 64 mm
On-Column Injection (d)
Syringe Needle Length 80 mm
Detector depth, insert until it touches the jet tip, pull back 1-2mm then tighten nut.
Note all the injector settings are the same for the Trace GC so you can download the manuals etc.
Chromcard; The method is the whole file inc your sample table, always save it to a new name (and directory) as it allows you to go back and re-process. I may still be the UK expert on Chromcard which IMO is a great GC integration package. Are you running the old 16 bit version or the 32 bit one ?