I´m analyzing Olanzapine, raw material, by USP 34 method. I´m doing Related compounds, and I´m having problems on stabilizing the gradient. The gradient begins to rise at 10 min on Mobile phase B for ten minutes, plateau from 20-25, and initial conditions at 27, and ends at 35. The problem is that is never the same shape of the base line, and I´m pretty sure I have stabilized it before each run, because I let it run for ten minutes longer after 35 minutes. I´m using a Luna C8, and the mobile phase A and B are different compositions of Acetonitrile and Sodium lauryl sulphate buffer. System temperature is 35 °C. Sometimes the gradient goes up to 200 mAU, and the next run goes to 15 mAU for example, randomly. Anyone knows what could it be happening?
Q. F. Ignacio Viera
Ion-pair reagents are notoriously slow to equilibrate and are also temperature sensitive. So:
- try equilibrating longer.
- tee a thermocouple into the flow stream just ahead of the column and monitor the mobile phase temperature to see if there is some correlation there.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Bing [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry