Pore Size: GE HiTrap Ion Exchange

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I neglected to filter my sample through a 0.45 micron filter, and I'm worried my sample might be stuck inside my GE HiTrap Q 1 ml anion exchange column. (I'm working with Vitamin D, which forms aggregates larger than 0.45 microns...)

Is there any way for me to clean my column, or run it backwards?

Would particles larger than 0.45 microns actually get stuck in the column, or flow through?
if i remember well we are talking of a something like 30u size resin in the column?

the rule of thumb is to filter 1/10 of the particle size because this is the expected size of the channels between the particles and not taking into account the frit type you have at the head of the column
i really do not remmeber the size of the frit.
The nominal particle size is 34 um. 1/10 of that is 3.4 um, or 3400 nm - in my case, that gives me plenty of leeway.

Why does the manual tell me to filter closer to 1/100 of the size if 1/10 is the rule of thumb?

Thanks!
Why does the manual tell me to filter closer to 1/100 of the size if 1/10 is the rule of thumb?
They may be using a very small porosity frit. My suggestion would be "when in doubt, follow the manual".
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry