Hi guys, What I have is listed:
1- safety shutdown before reaching the set point for the pressure and flow rate.
2- very slow flow rate elevating to the set point.
3- I was able to zero the PPE valve with the gas (H2) supply off and the column in the injection port is out.
4- as soon as I turn on the gas supply on and I put back the column in the injection port the flow rate is elevating by 0.2 mL/min from the zero. Same thing for the PPE pressure value.
5-gold seal and the washer underneath it has been changed. Also the ferrules in the column ends.
6- I was able to reach the set point for the flow and valve pressure by increasing manually 0.2 mL/min a time to 1 mL/min.

previously before this problem occur I noticed a decreasing in the flow rate from 1 mL/min to 0.991 mL/min at the end of the run when the oven temperature is 280 C and the pressure is high 17 psi.
thanks
Step 1: check the whole system for leaks, using a leak seeker.

Peter
Peter Apps
Just for you to consider. Long time ago, I used to use 5890, and I still remember that something happened just like you describe and the reason was a clog in the split valve. As I remember, it's an electronic control back pressure valve.
Or a clog in the weldment. Or a clog in the split VENT line; that includes the trap.
[quote="Consumer Products Guy"]Or a clog in the weldment. Or a clog in the split VENT line; that includes the trap.[Yes, It was the weldment, Thanks for the help]
5 posts Page 1 of 1

Who is online

In total there are 198 users online :: 1 registered, 0 hidden and 197 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 197 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry