Here are detailed of LC-MS user manual about solvents and Buffers
-------------------------------------------------------------------
Solvents :
ESI requires polar solvents. Non-polar solvents, however, can often be used successfully if a polar modifier is added. For example, toluene, a non-polar solvent, modified with 15% isopropyl alcohol can be used as a solvent for the ESI analysis of fullerenes in negative ion mode. The following table includes examples of other solvents that can be used for normal-phase chromatography when modifiers are added. For positive ionization, mixtures of acetonitrile/water, methanol/water, and isopropyl alcohol/water are most common but other mixtures can be used with success. Acetonitrile/water, isopropyl alcohol/water and n-propyl alcohol/water are good starting mixtures for negative ionization. API-electrospray sensitivity is best with either acetonitrile or methanol and water. Typically, the pH of the mobile phase is adjusted in order to cause the highest yield of ionization in the solution phase.
Partial list of solvents and their suitability for ESI
-----------------------------------------------
Commonly used Special cases
-----------------------------------------------
Water (<80%)Benzene,1
MethanolCarbon disulfide,1
EthanolCarbon tetrachloride
n-Propyl alcoholCyclohexane,1,2
Isopropyl alcoholHexane,1
t-Butyl alcoholLigroin,1
AcetonitrileMethylene chloride,1,2
AcetoneToluene,1,2
Tetrahydrofuran
Acetic acid
Formic acid
Chloroform
Formamide
-------------------------------------------------------------------------------------------
1,Requires modifier
2,Normal-phase chromatography
--------------------------------------------------------------------------------------------
Buffers :
Buffers are used for many reasons including:
Adjusting solution pH to support ion formation in solution (generally, positive analyte ions are formed more readily in acidic solutions and negative analyte ions are formed more readily in basic solutions). Ensure formation of specific desired adduct ions or prevent the formation of undesirable adducts Assist or optimize chromatography a buffer is added. Buffers that assist or optimize chromatography and those that do not hinder the electrospray process can be added before the separation. Buffers that interfere with the separation must be added post column. For most positive ion analysis of polar materials such as amino acids, peptides and proteins, the pH of the solution should be adjusted to a pH of 2 – 5. The addition of acetic acid at 0.1% to 0.2% is a good starting point. For positive ion analysis of pharmaceuticals, a solution of 0.015 % formic acid serves the same purpose and may have less chemical noise and smell than acetic acid. Some pharmaceutical compounds can be analyzed successfully in a neutral mobile phase. For example, benzodiazepines and opiates can be analyzed with a traditional mobile phase of acetonitrile and water. Buffers such as sodium acetate or potassium acetate (alkali metals) can be used to form adducts with the analytes that would otherwise not ionize in solution. Sugars and urea are two examples of chemicals that form sodium adducts that can be analyzed in positive ion mode. Other buffers, such as ammonium acetate and ammonium formate, are sometimes added to prevent undesired adduction of the analyte with sodium or potassium ions from endogenous sources. Buffers can be used to optimize chromatography. The addition of 50 micromolar ammonium acetate or ammonium formate is often used to increase chromatographic resolution of basic nitrogen containing compounds on reversed-phase silica columns. This improves the peak shape, thereby enhancing signal and improving sensitivity. The final solution (solvent + analyte) should be neutral to acidic for good positive ionization.
Buffers or other additives used to optimize chromatography can sometimes interfere with the ionization process. For example, TFA is almost always used for the chromatography of peptides and proteins. TFA enhances the chromatographic
resolution but may actually suppress ion formation. Post-separation addition of a weaker acid such as propionic acid can effectively counteract the TFA ion suppression problem.
When performing ESI standard buffers such as phosphate, borate, and sulfate buffers are non-volatile and form ion pairs in solution. To maximize ESI sensitivity, use buffers that are volatile and do not form ion pairs. Adjust the pH with buffers, formic acid, acetic acid, and ammonium hydroxide or triethylamine. Typical pH for positive ion is neutral to pH 2 and for negative mode typical pH is neutral to pH 10. For ion pair separations, use additives such as heptafluoro butyric acid or tetraethylammonium hydroxide or tetrabutylammonium hydroxide.