Have somebody tried to detec lactose by HPLC-MS/MS or HPLC-DAD? I am trying to detect lactose in milk with LOD below 0.03 %.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
If I'm interpreting this correctly, 0.03% (w/v) would be equal to 0.3 mg/mL = 0.3 micrograms/microliter. Assuming a typical injection volume of, say 10 microliters, that would be 3 micrograms on-column per injection. That should be well within the working range of something like RI detection. Barring issues with interferences, why would you need to use LC-MS/MS?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
This is easily achievable on standard LC equipment with ELSD detection.
http://www.waters.com/waters/library.ht ... d=10179824
Download the applications book and it's application WA64095
http://www.waters.com/waters/library.ht ... d=10179824
Download the applications book and it's application WA64095
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 76
- Joined: Tue Oct 05, 2004 10:41 am
Thanks a lot,
I will let you know how it goes!
I will let you know how it goes!
-
- Posts: 20
- Joined: Thu Sep 17, 2009 9:55 pm
Can anybody share the mass spectrum of lactose by lc ms/ms?
I am working on lactulose by lc/ms/ms. Lactulose have M.W. 342. For Posivtive ESI, I could not find the M+1 ion, 343. I found m/z 360. Can anybody help to explain this?
Thanks
Jen
I am working on lactulose by lc/ms/ms. Lactulose have M.W. 342. For Posivtive ESI, I could not find the M+1 ion, 343. I found m/z 360. Can anybody help to explain this?
Thanks
Jen
-
- Posts: 70
- Joined: Fri Jan 02, 2009 12:53 pm
looks like ammonium adduct, [M+NH4]+. what is your eluent composition?Can anybody share the mass spectrum of lactose by lc ms/ms?
I am working on lactulose by lc/ms/ms. Lactulose have M.W. 342. For Posivtive ESI, I could not find the M+1 ion, 343. I found m/z 360. Can anybody help to explain this?
Thanks
Jen
-
- Posts: 199
- Joined: Tue Feb 08, 2005 5:27 pm
Hola Carol.
You can analyze lactose by LC MS; but sometime this could be too much?? depending of what your requirements are. lactose can be analyzed by simple UV/Vis spectrofotometry with an enzynatic method; faster and easier but maybe not too much precise.
Other way is use Elechtrochemical detector.
You can analyze lactose by LC MS; but sometime this could be too much?? depending of what your requirements are. lactose can be analyzed by simple UV/Vis spectrofotometry with an enzynatic method; faster and easier but maybe not too much precise.
Other way is use Elechtrochemical detector.
-
- Posts: 75
- Joined: Fri Oct 07, 2011 10:16 am
Jen,Can anybody share the mass spectrum of lactose by lc ms/ms?
I am working on lactulose by lc/ms/ms. Lactulose have M.W. 342. For Posivtive ESI, I could not find the M+1 ion, 343. I found m/z 360. Can anybody help to explain this?
Thanks
Jen
look at the following for ESI -ive MS/MS
683.22 [2M-H], 341.11 [M-H], 281.09, 179.05, 161.05, 101.02
when I use the CID energy 15 eV, the following ratio was obtained
683.22 [2M-H], 31%
341.11 [M-H], 100%
281.09, 6%
179.05, 15%
161.05, 30%
101.02, 1%
you can get another ion with High energy, such as 221, 143 and 119
(Lactose C12H22O11 MW= 342.115663, [M-H] = 341.107838, [2M-H] = 683.224049)
This is obtained from my system only, I'm not sure it same as others.
Jetjamnong
Jetjamnong
-
- Posts: 20
- Joined: Thu Sep 17, 2009 9:55 pm
Hi Jetjamnong
You are using the ESI negavitve ion, I am using ESI posivtive ion. Is there any advantage of Negative ion over Positive ion in this application?
Thanks
Zhen
You are using the ESI negavitve ion, I am using ESI posivtive ion. Is there any advantage of Negative ion over Positive ion in this application?
Thanks
Zhen
-
- Posts: 175
- Joined: Thu May 19, 2011 3:52 pm
Jen,
Most people do quantitation of sugars in negative mode where you are generating the M-1 ion.
Some people try to generate positive adducts and quantitate that but quantitation of an adduct
can be unreliable. As others have pointed out there are more linear and easier to use detectors
for sugars such as RI and ELSD.
Most people do quantitation of sugars in negative mode where you are generating the M-1 ion.
Some people try to generate positive adducts and quantitate that but quantitation of an adduct
can be unreliable. As others have pointed out there are more linear and easier to use detectors
for sugars such as RI and ELSD.
-
- Posts: 20
- Joined: Thu Sep 17, 2009 9:55 pm
Hi Jetjamnong
Thanks for the information, I will try the negative mode.
Zhen
Thanks for the information, I will try the negative mode.
Zhen
11 posts
Page 1 of 1
Who is online
In total there are 40 users online :: 3 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 37 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 37 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science