We have been working on USP method for one of my products.The details are as follows.

Buffer: Dissolve 0.5 g of tetrabutylammonium hydrogen sulfate and 1.0 g of monobasic potassium phosphate in water, add 2 mL of phosphoric acid, dilute with water to 1000 mL, mix, and pass through a 0.45-µm filter.

Mobile phase—Prepare a degassed mixture of Buffer solution and methanol (18:7).

Wave length:UV 293 nm.Flow:1.2 mL/min
Column:L11-250X4.6 mm-5 micron.

We have observed very high noise using above conditions with one brand column of L11 Packing.However ,no noise was observed when using other brands of L11.Is it possible that a specific column contribute to more noise killing sensitivity of the method?Please advice.
L11 is phenyl boded phase. Yes, some brands bleed more than others, depending on the bonding chemistry and the purity of the underlying silica.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry