Hello all,

after starting the run on schimadzu SCL 10Avp hplc i got an error as SIL: NO 6PV VALVE HOME. i got this problem first time. what shall i do please let me know if anyone has the right answer will be appreciated.

zulfaquar
Your autoinjector can't properly actuate the 6 port valve.
This error is usually caused by wear over time resulting in polishing of the mated surfaces. Once they are polished enough to allow for significantly more contact than with new components, the actuating motor can no longer overcome the friction between them, causing the error.
At the very least, a qualified person needs to open up the 6 port valve and inspect the rotor and stator. It would probably be best to go ahead and replace those parts.
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry