Page 1 of 1

HPLC Error!

Posted: Sat Oct 15, 2011 2:02 pm
by arfi02
Hello all,

after starting the run on schimadzu SCL 10Avp hplc i got an error as SIL: NO 6PV VALVE HOME. i got this problem first time. what shall i do please let me know if anyone has the right answer will be appreciated.

zulfaquar

Re: HPLC Error!

Posted: Mon Oct 17, 2011 3:35 pm
by DR
Your autoinjector can't properly actuate the 6 port valve.
This error is usually caused by wear over time resulting in polishing of the mated surfaces. Once they are polished enough to allow for significantly more contact than with new components, the actuating motor can no longer overcome the friction between them, causing the error.