Advertisement

Resolution?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
hi,
i'm running sunscreen which required me to achieve resolution at least 2.5between Avobenzone and Octinoxate peak. the mobile phase used r 60% 0.1% Aqueous Phosphoric Acid, and 40% Methanol, Thermo Scientific Hypersil Gold, 5micron 3.0 mmx150 mm, 0.4ml/min,254nm. i'm facing a problem to separate those two actives...can sumone suggest what kind of column or Mobile phase is suitable to separate both actives?? :
Try to do with Thermo Scientific application.
http://www.analiticaweb.com.br/newslett ... orbers.pdf
For summary,
Column : Hypersil Gold, 1.9 um, 100 x 2.1 mm
Mobile Phase A : Water, B : Acetonitrile
Isocratic : 40:60
Flowrate : 1.0 ml/min
Detector : PDA Scan 200-500 nm, Discrete 313 nm
Column Temp. : 45 degree C

Best Wishes
Jetjamnong
Jetjamnong
2 posts Page 1 of 1

Who is online

In total there are 283 users online :: 0 registered, 0 hidden and 283 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 283 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry