Poor Reproducibility of readings from the IC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi All,

I am having trouble obtaining the same readings for the same samples being run for cations on the IC. The readings tend to increase as more samples are run. Does anyone know of why this is happening?

Thanks :)
Can you be a *bit* more specific?
What instrument (manufacturer/model)?
What column / mobile phase?
What kind of suppressor?
Are the retention times changing or only the peak areas?
How much variation are you seeing?
How suddenly did the problem begin?
Did the onset of the problem coincide with some specific event (instrument service? new batch of mobile phase?)?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry