Posted: Wed Oct 12, 2011 12:48 pm
good day!
when I think of quenching, I think of reduced absorption, though in this case im witnessing increased absorption, or perhaps not...
I calibrate my nitric oxide measurement with N2 diluted known bottle concentrations (<2000ppm NO)
when I measure diesel combustion exhaust, the sample is infused with H2O and CO2 (eg. 8%, and 14% respectively)
experimentally, I vaporize water and mix with known NO and CO2, and see that the NO measurement is significantly higher than my known value, proportionally to the concentrations of H2O and CO2
I can solve for this error linearly and my FTIR measurement is correct experimentally and when measuring exhaust
but I cant explain the phenomena I am seeing, any ideas?
thank you kindly!