peak width and detector senstivity
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
I always have one question. As detector become more sensitive, will peak become fatter? I always observe that MS peak is much fatter than UV peak. if this is the case, how can we calculate intrinsic plate number for column?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
So long as you are within the linear range of the detector, the answer is "No". It can look that way if a large part of the peak is off-scale, but peak width (measured as either baseline width or width-at-half-height) remains constant. If you go above the linear range of the detector, then width measurements will become unreliable.As detector become more sensitive, will peak become fatter?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 158
- Joined: Wed Mar 05, 2008 2:56 pm
Hi
Check your tubing and connections between the detector and the MS, you maybe getting some band broadening, hence the wider MS peaks in relation to the UV.
Check your tubing and connections between the detector and the MS, you maybe getting some band broadening, hence the wider MS peaks in relation to the UV.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thanks. this makes sense.
I was totally off when I posted this thread.
I was totally off when I posted this thread.
So long as you are within the linear range of the detector, the answer is "No". It can look that way if a large part of the peak is off-scale, but peak width (measured as either baseline width or width-at-half-height) remains constant. If you go above the linear range of the detector, then width measurements will become unreliable.As detector become more sensitive, will peak become fatter?
-
- Posts: 18
- Joined: Fri May 27, 2011 2:55 am
I am going to assume the simplest interpretation of your question. That you are comparing sensitivity [really limit of detection] between UV and MS and the MS peaks are also broader. As LCBob suggested individual fittings with large void volumes are most frequently to blame. Another major culprit can be the UV detector cell especially if it is older generation. Next, the exit tubing from the UV detector may be of much large ID. A final area of broadening is frequently in the valved inlet of the MS itself.
If you suspect the UV cell, either remove it or perform a split prior to the UV detector as a test to see if peak width reduces for the MS. As a general rule keep all tubing < 0.010" id and if possible 0.007" id even up to the mass spec.
If you suspect the UV cell, either remove it or perform a split prior to the UV detector as a test to see if peak width reduces for the MS. As a general rule keep all tubing < 0.010" id and if possible 0.007" id even up to the mass spec.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Exit tubing on a cell contributing to peak broadening?
-
- Posts: 512
- Joined: Tue Jul 19, 2011 9:26 pm
Peak broadening on the MS when it's connected after the UV - yesExit tubing on a cell contributing to peak broadening?
7 posts
Page 1 of 1
Who is online
In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 27 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 27 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science