separation of hydrocortisone acetate n cortisone acetate

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
hi,
i'm using bp method for hydrocortisone acetate related substances which required me to achieve resolution at least 4.2 between hydrocortisone acetate n cortisone acetate peak. the mobile phase used r 40% ACN, C18,250x4.6mm,5micron, 1ml/min,254nm. i'm facing a problem to separate those two drugs...can sumone suggest what kind of column r suitable to separate both drugs?? :
Since you are running a compendial method, your best bet would be to find the original monograph to determine the column that they used in the first place. If you cannot get that exact column, then you can search the PQRO database on the USP website (http://www.usp.org/USPNF/columnsDB.html) to find the closest equivalent.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry