- Posts: 656
- Joined: Tue Jul 05, 2005 7:45 am
Direct comparison of Acquity BEH C18 (1.7 µm) and Kinetex C18 (1.7 µm):
Analysed the same day on the same instrument, using the same sample and mobile phase:
Instrument: Waters Acquity
Mobile phase: 20% acetonitrile in 5 mM acetate buffer pH 4.5 (isocratic)
Sample: 20 µl of a 100 µg/ml solution (in-house peptide)
Flow: 0.5 ml/min
Temp: 70C
Column dimensions: 2.1x150 mm
Pressure:
Acquity column: 9400 psi
Kinetex column: 8400 psi
(I am surprised that the pressure difference is not bigger. I thought the core-shell technology would make a larger difference in back pressure!)
Retention time
Acquity column: 18 minutes
Kinetex column: 14 minutes
(More retention on the Waters column, maybe not so surprising)
Tailing factor
Acquity column: 1.7
Kinetex column: 1.4
(clearly better symmetry on the Kinetex column)
Plate count
Acquity column: 12.000 plates
Kinetex column: 16.000 plates
(I am a bit shocked that the Kinetex column gave so much better efficiency. I may have overloaded the Acquity column, giving a lower plate count and more assymmetry)
Conclusion: The Kinetex and Acquity columns are probably comparable. The Kinetex column appears to have a better loadability than the Acquity column. The back pressure was quite similar of the two columns, which means that the Kinetex column (1.7 µm) must be used on a UPLC or UHPLC system.
Feel free to comment!