Hi all, could anyone tell me which kind of tests should I try if I intend to perform an analytical method validated in an HPLC with a standard DAD detector to another HPLC with a more sensitive DAD detector?
Thank you in advance.
If the detector is more sensitive than the original one, I wouldn't expect big problems. Signal/Noise ratio might be checked, if you're working with low concentrations, but it should be better with the more sensitive detector.
One thing often forgotten might be to check which detector settings (sampling rate!) are appropriate with the newer one...
Honestly, I wouldn't expect a formal revalidation to be necessary...
I agree with HPLCaddict: so long as you pass system suitability on the new system, revalidation should not be necessary.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I agree with HPLCaddict: so long as you pass system suitability on the new system, revalidation should not be necessary.
That is true if the SST contains a "sensitivity criterion". If it hasn't, compare the S/N of low concentration standard.
On the other side, if you work at higher absorption (500-1500mAU), it would be sensible to check linearity.

Alex
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry