Good afternoon,
I am having some trouble in one current method we use to quantify some anti-infeccious drugs from plasma samples. We use an Acetonitrile:Phosfate Buffer gradient by 25 minutes at 215 nm, witch results in a perfect separation of peaks. Our system is an Elite LaChrom and the coullum is an Waters Sunfire. The method had a normal drift of baseline, as the acetonitrile proportion rises up, at max 0,005 mUA, witch we accept as rasonable.

Last week we exange the collum to try a new method on an Waters Atlantis, and when we put the Sunfire back in the system and calibrate the anti-infeccious drugs method, the drift get up at max of 0,2 mUA, witch is incompatible with our LLOQ.

What could be causing this?
If you have not already done it, prepare fresh mobile phase (particularly the phosphate buffer). Microbial growth in aqueous buffers can give rise to that sort of drift. While you're at it, check the needle rinse solution in the autosampler.

Try purging the column and system for a few hours with acetonitrile and then run a few blank gradients. It's possible that something may have bled off the other column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 34 users online :: 3 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry