thanks for your answer.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Got it covered, carls, but it is a good point to re-iterate - if you don't calibrate your VWD per Agilent's specs before the test sample, the results mean nothing (according to Agilent, that is...). And if you calibrate after the test sample, you still need to run the test sample again to confirm the calibration. Although, performing the calibration technically only involves the deuterium lines and holmium oxide filter, according to Agilent. The caffeine wavelength test might be a nice external test, but I don't think it's one you would be able to hold Agilent to, unless you got an engineer to repeat it on-site with your instrument. Are you required to run this test for some regulatory agency requirement, or is this only for your own personal satisfaction?Also from the pdf - "Wavelength accuracy - ± 1 nm Self-calibration with deuterium lines, verification with holmium oxide filter". If you are able to run this VWD scan and your UV-max doesn't match what is described on the certificate you have (within ± 1 nm, that is), you might want to run the VWD calibration under the Diagnostics section of the Chemstation software.