pump seal failure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
ı have been working in agilent 1200 systems. my method is containing reversed phase solvents.

ı change my pump seal every 3 or 5 mounths. ı think this time span is lower life for pump seal. ıt reduces a big cost for our. my shut down method containing a acetonitrille. ı red a article about pump seal ' higher acn value can reduce pump seal because acn to wet the pump seal surface than meoh'

what is the possible cause this problem?

thanks every body.
Check with Agilent regarding their recommendations for seals. In my experience, premature seal failure is more often associated with leaving buffer-containing mobile phases in the pump at shut down (as opposed to flushing the buffers out of the pump), so a shut-down with ACN is generally a good idea.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry