We have a fairly new (purchased 2008) Beckman System Gold (model 126) HPLC, on which we run (for the most part) 0.1M phosphate buffer (pH 6) on pump A and HPLC-grade MeOH pump B (all solvents pre-filtered). We have been changing the inlet + outlet check filters every 3-5 months, because of occasional high back-pressure during the method run; however, recently we noticed that the pump pressure would gradually increase at the end of the run to at least 200 bar (2.9 kPsi) and simultaneously flow from pump A would stop. So we tried introducing a water wash at the end of the method (which is when we would usually see this problem), and that seemed to help matters, but only in that it decreased the frequency with which this problem occurred. We changed out the high-pressure seal a couple of times as well, and after the last change we have noticed that our actual flow does not match the set flow-rate. We've been in touch with Beckman and they suggested we do a sequential high-pressure check to rule out the inlet and outlet check valve cartridges, but we still have not been able to find a cause or solution for our problem. ANY input at this point would be helpful!
Thanks.