liner issues on GC-FID

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all, I am running a GC-FID analyzing volatile fatty acids. It is direct aqueous injections on a DB-FFAP column. Because of the nature of the samples (supernatant from digested sludge water) I have to change the liner routinely. The problem is that when the new liner (wool packed) is installed I have to "condition" the liner by running a series of low to high standards followed by 30 blank injections. If I don't do this I have a very noisy baseline and the areas of my lowest standard (12mg/l) will not meet. However, sometimes even after this conditioning step my blank sample (previously analysed as below LOD at 2mg/l) still reads above LOD. If anyone has any suggestions on how to avoid or rectify this it would be gladly appreciated.
Perhaps the use of silanized glass wool can help you. See Supelco catalog for this.
2 posts Page 1 of 1

Who is online

In total there are 51 users online :: 0 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry