SPME, VOC, sampling time and solvent cut time

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I am a new user of SPME. I will use SPME to analyze VOC. I check the paper in Supelco for the detail about inj. The condition paper suggested is: “splitless, 230C, closed 3 min.” I am not sure about “closed 3 min”. Does it refer to the interval between the time of SPME fiber expose in the injection port and the time at which the split flow vent valve is opened?

If my understanding is right, I have to setup “sampling time” in my Shimadzu (QP2010+) software GCMSsolution for 3 min. But it causes a problem that the “solvent cut time” has to be at least 3 min. as Shimadzu software GCMSsolution requires that “solvent cut time” is not shorter than “sampling time”. However, for VOCs the retention time is less than 3 min. Can anyone have any suggestion for me to have 3 min. sampling time and 0-0,5 min solvent cut time?

Thanks a lot.
We do VOC's by SPME and use a cryo module . You may need one or get a lomger column. Beware that if you do not use cryo your peaks will be quite broad.
2 posts Page 1 of 1

Who is online

In total there are 52 users online :: 0 registered, 0 hidden and 52 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 52 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry