Trouble with upper and lower limits of linearity on a GC

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I am having issues with some re-validations on some of our older instruments. I have five GC/MSs that are being used to test for drugs of abuse (THC specifically) Two are 6890 / 5975 combos and three are 6890 / 5973 combos. All are running the same Chemstation revision. On the 5973s I can get the required response at the upper limit of linearity but not the lower limit, or I can make some adjustments and have the needed response at the lower end, but not the high end. The 5975s run the same samples just fine. We have eliminated all of the easy fixes and are rather stumped at this point. Any thoughts?
Have you tried using the larger diameter drawout plate on the source? This was introduced to help with linearity issues that some 5973s were having.
Have you tried using the larger diameter drawout plate on the source? This was introduced to help with linearity issues that some 5973s were having.
I didn't know such a thing existed. I'll check into it.
3 posts Page 1 of 1

Who is online

In total there are 50 users online :: 0 registered, 0 hidden and 50 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 50 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry