sutible coulmn for retain benzene

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
What type of stationery phase could be most interaction with benzene? I want to analysis some impurities in benzene by internal standard method. I planned connect to column together which first column retain benzene while my desired impurities (mostly light dienes) pass without retain. After that I can use second column to separate desired impurities. By this way benzene peak will be appeared near end of chromatogram.
Now, I am looking for column which has most interaction with benzene. I tried CP-wax column as polar phase but it can not be able to shift benzene peak enough.
Would you please suggest another type.
The TCEP phase column is a traditional favorite, it shifts benzene a carbon number later !0.5 than the Wax (9.5) phase.

There is a new phase sold by Supelco that has only been on the market one or two years which is reported to shift benzene another carbon number (11.5).

It is the only new development that improves performance over the TCEP in the last 50 years !

best wishes,

Rod
Use old good phenyl based columns (similar with similar is the rule).
Use good brand, use longer, narrower and higher film thickness, at least 1 or may be even 3 um film. The ultimate solution will be GCGC, but 100m column is definitely cheaper than GCxGC.
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
Alexandre

mr3573 wrote:

"I want to analysis some impurities in benzene by internal standard method. I planned connect to column together which first column retain benzene while my desired impurities (mostly light dienes) pass without retain. "

He wants to separate alkanes and mono- and di- enes from the aromatic benzene. Your suggestion would be an exercise in futility to accomplish his goal.

One of the new ionic liquids will perform the separation he desires while he can then use a second column, usually a boiling point column to separate and identify the impurity components.

Traditionally the TCEP phase was used to perform the first separation.

best wishes,

Rod
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry