by
Karen01 » Sun Aug 07, 2011 4:16 pm
Thanks for the input. It really helps me understand.
I should have mentioned this is a headspace run. Is it not advisable to run splitless with a headspace?
No IMO if you have loop headspace sampler and you have the instrument setup correctly for headspace.
First you should have the septum purge turned off and the instrument plumed (or flows set) so that all (or most depending on how plumbed) of the carrier comes into the inlet through the headspace transfer line. I've had 5890s,6890, and 7820s (well can't turn off septum purge there) set up that way for headspace. That minimizes sample dilution.
With that setup empirically I found that higher splits give shaper peaks (so better chromatography/separations) and sometimes even more sensitivity.
I think the reason has to do with band width.
With a loop headspace sampler if your column flow is 1 ml/min and you have 1 mL loop, as 1st approximation it will take 1 min to flush the sample from the loop and so the the analyte will be deposited on the column over a 1 minute. If it migrates at all at the initial column temperature you wind up with broader peaks, lower peaks with potentially less resolution between peaks.
The higher the split the the less material (which helps separations also), but the narrower the band on the column.
Does that makes sense?
I know you are likely running a validated method you can't change... but if you have the freedom to play a bit and your instrument is setup correctly for headspace, try injecting your low standard (or dilute it to the bottom of your quantitation range) at 5:1 and 10:1 splits and see what the injections look like.
HTH
- Karen