We are in the process of porting a Foot-and-Mouth Disease virus quantitation at 260 nm from SEC FPLC with Sephacryl S400 in an GE Äkta UPC-10 to HPLC in a TSKGel G4000PWXL column. With the low pressure Sephacryl columns we had excelent linearity over almost 2 orders of magnitude (see Vaccine 29 (2011) 7182– 7187). We bought an Agilent Infinity 1200 with a VWL detector and automatic injector as a dedicated equipment for this purpose. During early testing we had similar results as FPLC, but when we attempted a area under the peak vs. dilution test we got a catastrophic loss of linearity. Values of area decline steadily from the expected value with increased dilution, e.g. at a 1:6 dilution of the standard (1 volume standard + 5 volumes diluent), instead of the expected 0.1667 of the area of pure standard we get 0.079, and the difference gets much worse for higher dilutions.
We are also observing a systematic difference between replicated injections of the same vial, with the second injection producing 20 to 30% less area than the first one. This behaviour was observed in sextuplicate vials, in each one the second injection was lower. The difference was statistically very significant in a t-test and was observed at two different concentration levels (1:15 and 1:72 of the pure standard).
I have no clue at what is going on, please help me!